Issues with electro-magnetic noise #11
-
Dear Oliver, Originally posted by @Matthias32 in #10 (comment) |
Beta Was this translation helpful? Give feedback.
Replies: 15 comments
-
The second image looks quite good. 1-2 ms pulse width as expected on the downward-facing part. The regular artifact in the first image looks like EM noise or stray light - it seems to be close to 50 Hz. Could you post a picture of your whole measurement setup and also show the alpha radiation window if you made one? |
Beta Was this translation helpful? Give feedback.
-
This was the setup. The alpa radiation windows is just a hole in the bottom of the case. |
Beta Was this translation helpful? Give feedback.
-
From this picture, it is difficult to give advice. I would need to see how the case itself looks like, including the radiation window etc. Honestly, an aluminum foil applied like this will do not much. It is usually not even good in blocking light due to tiny cracks or small openings/folds close to the table surface. And for shielding low-frequency EM radiation - which is absolutely required here - it also does not work since aluminum always builds up a non-conductive oxide layer. It simply cannot be used to form a proper Faraday cage - that function should be already taken care of by the metallic enclosure. |
Beta Was this translation helpful? Give feedback.
-
Thank you a lot. This is the detector below the aluminium foil. |
Beta Was this translation helpful? Give feedback.
-
Hi Matthias, Even with a metal foil glued on top, the diode should be always facing down and away from sources of light. EDIT: do you have the 9V battery outside of the case? The short adapter cable on one of the pictures may suggest that... Keep me posted. |
Beta Was this translation helpful? Give feedback.
-
Hi Oliver, Where I can get these thin metal foils from? Or how should I disassemble the foil capacitors to get the foil? |
Beta Was this translation helpful? Give feedback.
-
Many types of foil or film capacitors should work as long as there is no electrolytic liquid chemistry inside - non-polar film capacitors should be mostly dry inside. I used those large old types (the size of a thumbnail) from an old electronic device (pre 1980 I would say) which seems to have a kind of "ceramic" layer of paint around them (the ones that look more like pills, with rather organic/non-flat surfaces). I simply cut it open with strong pliers. If I find the pictures again, I'll post one. Did you have some luck with thick black cloth on top of everything? That should really work as long as there is not even the smallest gap between the table surface and cloth allowing light leaking inside. |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.
-
Hi Matthias, thanks for your many pictures and explanations, it makes it much easier to follow! I guess what you observe in 2) and probably as well 3) is leakage current coming from the main power (you'd short circuit the tiny leakage current on your mains' earth connection by touching the case). The ground/shield connection of all BNC sockets on mains-powered oscilloscopes is usually directly connected to the earth of the wall outlet (via the so called "Schuko" in Germany). If your type of wall socket allows that, you could try turning the oscilloscope power connector by 180 degrees and with some luck the leakage current could drop depending on the internals of the power supply inside the oscilloscope. But the source of leakage current on the earth connection could be as well some other high-power device on the same mains network like a fridge or kitchen stove in other rooms (up to 2 mA of AC voltage are allowed to be disposed by a power-hungry device on earth connections and per device if I remember correctly). Now the question is: how does the leakage current couple into the detector circuit? I had assembled an alpha detector similar to your approach about a year ago. Unfortunately I gave it away, so I can't confirm, but I think that I had isolated the diode with some tape and then pressed it into the hole of the aluminium case. The purpose of the tape would be that aluminium case and diode are electrically well separated. The type of tape could even play a role with different permeability and electrostatic properties... could be worth trying out few different ones if the first one doesn't work right away. What is your alpha source by the way? It is shielded/encapsulated well, the emitted alphas could loose already most of their energy before they reach the detector, irrespectively of the additional capacitor foil. Let me know how it goes. |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.
-
ok, at least we can address why you don't see alpha pulses: Uranium glass is usually not a noticeable source of alpha particles. It should be in theory but in practice, the uranium concentration is so low, that you'd need to be very lucky to find a spot right on top of the surface of the glass where atoms decay and related alphas are not already absorbed by the glass. So far I have only found thoriated glass from flea markets, like these old camera lenses, to actually emit alpha particles (besides a lot more of beta and gamma, because those lenses can be very active): https://camerapedia.fandom.com/wiki/Radioactive_lenses Besides old smoke detectors with Am-241 sources inside (sometimes sold as "ionization chamber", cough Aliexpress cough), an easier to find and harmless kind of everyday alpha source is old ceramics featuring uranium glaze: https://www.google.com/search?q=uranium+glaze. A less intensive alpha source, but freely available everywhere, can be collected from radon in the air of any regular room: https://iopscience.iop.org/article/10.1088/0031-9120/32/2/016/meta Regarding the noise problem: If I understand correctly, having the case around the circuit does not make much of a difference. From this, I would conclude that the ground connection between the circuit and the case is not sufficient. You should have at least a short GND wire coming from PCB and soldered to the BNC connector. If that one is firmly mounted in the case it should be fine. In addition, I guess you are using at least one or two metal screws that hold the PCB in place through either the three large PCB holes or the two exposed PCB edges that are all grounded. Looking at the uranium marble and foam, I hope you don't expect that it is blocking the light well enough. The alpha spectrometer circuit is not yet as sensitive as a single photon-detector but really rather close to that. The light coming through the foam and marble would be by far too much if either the diode or the whole detector setup is not kept in complete darkness. Could you maybe record a short video using the battery-powered oscilloscope, ideally in another building/place and demo the problems? |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.
-
Thanks for the video! :-) The T0 metal case of the diode itself does not really contribute to EM shielding. The whole first amplifier stage is extremely sensitive to charges and stray capacitances. Regrading the frequency properties, there are different factors at play. EM noise can couple in different ways. I'd recommend to check out the literature on EMI for product development (German 'EMV'), it's a fascinating but really complex topic... |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.
-
Congratulations! Looking at your upper video again, it could be dominated by fluorescent/LED room light - that may be bouncing back from the table surface as you approach it - and causing the strong 50 Hz waveforms. Generally, one layer of black insulation tape is often not enough for blocking all the light in a well-lit room and depending on the tape. The main difficulty is that "mains clocked" light and other mains powered EM coupling may induce similar sine waves such that both may be contributing at any time if not shielded enough individually. |
Beta Was this translation helpful? Give feedback.
Congratulations!
The pulse amplitude looks well bigger than the -8 mV threshold which, if measured at regular 1 Megaohm input impedance and 1:1 input attenuation, renders it a full-fledged alpha particle pulse.
c.f. https://github.com/ozel/DIY_particle_detector/wiki/Oscilloscope-Measurements
Looking at your upper video again, it could be dominated by fluorescent/LED room light - that may be bouncing back from the table surface as you approach it - and causing the strong 50 Hz waveforms. Generally, one layer of black insulation tape is often not enough for blocking all the light in a well-lit room and depending on the tape.
The main difficulty is that "mains clocked" light and other mains …